Wednesday, July 23, 2008

2D EXSY

The 2D EXchange SpectroscopY (EXSY) technique is exactly the same as the same as the 2D NOESY technique. The pulse sequences are identical. The method provides off-diagonal responses for spins which exchange slowly with one another (either conformationally or chemically) and also between spins with NOE's. The EXSY method is useful for showing exchange when the rate of the exchange is greater than or of the same order as the T1 relaxation rate (1/T1) but less than the frequency difference between the two spins (in the absence of exchange). Depending on the experimental conditions, the responses due to exchange are often much more intense than those due to NOE's. The figure below shows the 300 MHz 2D 1H EXSY spectrum of N,N-dimethylacetamide at room temperature. At this temperature, the molecule exhibits slow rotation about the (CH3)2N - C bond such that both methyl groups exchange with one another rotationally yet are distinct in the spectrum. This is evident by the cross peaks in the spectrum between the two methyl groups on the nitrogen. Note that the off diagonal peaks are of the same phase as the diagonal peaks.

7 comments:

Unknown said...

Hello Glenn

Firstly congratulations for your very usuful and great blog. Regarding Exsy-2d experiments you said that sequence is identical to NOESy-2d. There is a kind of filter for eliminate NOE peaks?

thanks

Ricardo

Glenn Facey said...

Ricardo,

Thank you for your question. AS far as I know there is no "filter" to discrimminate between NOE's and exchange peaks. This is not usually too much of a problem for small molecules as the exchange peaks are much stronger than the NOE peaks and of opposite phase.

Glenn

Craig said...

Hello Glenn

Would I be correct then in saying that it is possible to see both positive and negative signals (opposite phase and in-phase wrt diagonal) for the same proton signals arising from NOE contacts and simultaneous slow conformational exchange?

I have such a spectrum on my desk at the moment - the first example I've seen of this behaviour but it seems to be the only explanation.

Thanks,
Craig.

PS for Ricardo - there's supposed to be something called CLEAN-EXSY - here's a snippet from the Bruker documentation "Modified EXSY pulse sequences have been proposed to remove NOE effects (CLEAN-EXSY experiment) and to study protein hydration (89JACS1871)".

Glenn Facey said...

Hi Craig,

Thank you for the comment. Yes, I have seen both NOESY and EXSY peaks (of opposite phase) for the same protons. See an example here

http://u-of-o-nmr-facility.blogspot.ca/2012/01/sorting-out-noes-for-exchanging.html

Also thanks for the info CLEAN-EXSY.

Glenn

Glenn Facey said...

Bhagyawanti,
For a 13C-15N shift correlation experiment via J coupling you would require a sample double labelled in 13C and 15N. You could modify a standard hetcor experiment such as 'hxcoqf' to accommodate 15N rather than 1H and then add 1H decoupling.
Glenn

Pavel said...

Thank you for this blog - it's of great help! Will I see EXSY peaks in a 1H-1H 2D ROESY (roesyphpr) experiment as well? As a follow-up question: if I see no cross peaks between two potentially dynamic (NH) protons in ROESY, can I rule out a chemical exchange between them? Thanks, Pavel

Glenn Facey said...

Pavel,
Exchange peaks can appear in a ROESY spectrum just as they do in a NOESY spectrum. If exchange peaks are not present n the ROESY spectrum, it does not necessarily mean that the NH resonances are not exchanging with one another. It could be that the mixing time for the ROESY is not long enough or perhaps too long to see the exchange cross peaks. This depends on the rate of exchange. Things are a bit more complicated for samples in water where the NH’s are likely to exchange with the water as well.

- Glenn