tag:blogger.com,1999:blog-3300702123878659843.post1373503834730711654..comments2024-03-26T05:25:50.831-04:00Comments on University of Ottawa NMR Facility Blog: How Can I Get a Quantitative 13C NMR Spectrum?Glenn Faceyhttp://www.blogger.com/profile/05146575170575279335noreply@blogger.comBlogger14125tag:blogger.com,1999:blog-3300702123878659843.post-1769260589723973682021-03-15T21:14:01.370-04:002021-03-15T21:14:01.370-04:00Hi Josh,
Yes, shorter high power pulses will help ...Hi Josh,<br />Yes, shorter high power pulses will help with providing a larger bandwidth for uniform excitation. I would typically use 30 degree pulses. I think that the spectral width is less of an issue on modern instruments using digital filters and oversampled data provided there are no “smile” or “frown” artifacts near the edges of the spectrum. I would also add that the FID should be zero filled sufficiently to provide adequate digital resolution in the Fourier transformed spectrum.<br /><br />GlennGlenn Faceyhttps://www.blogger.com/profile/05146575170575279335noreply@blogger.comtag:blogger.com,1999:blog-3300702123878659843.post-49718505570024803422021-03-15T10:59:26.557-04:002021-03-15T10:59:26.557-04:00Hi Glenn. Great post. I have a couple of questions...Hi Glenn. Great post. I have a couple of questions:<br />1) Do you recommend using a shallower tip angle, which would enable shortening the relaxation delay? I'd think that a shorter pulse width would also improve the uniformity of the excitation bandwidth as a bonus.<br />2) Would you recommend increasing the spectral width? One of my users pointed to this article (JACS v.121, p.9455 (1999)) that recommends acquiring a 400 ppm wide 13C spectrum, and I'm not sure why. I suspect that it was for some kind of ad hoc oversampling method, which wouldn't be necessary on current instruments. If anything, I'd think that using an over-wide SW on a modern spectrometer would only enhance problems with phasing and baseline correction.<br /><br />I'll strongly consider the paramagnetic agent strategy, which will depend on my users' comfort level with introducing this material into their sample.<br /><br />Thanks! - Josh<br />Josh Kurutzhttps://voices.uchicago.edu/chemnmr/noreply@blogger.comtag:blogger.com,1999:blog-3300702123878659843.post-57202207320296616862018-04-04T11:45:49.932-04:002018-04-04T11:45:49.932-04:00Bo Zhang,
Yes of course the addition of a paramag...Bo Zhang,<br /><br />Yes of course the addition of a paramagnetic material will also shorten T2 and as a result broaden out the NMR lines. One would have to experiment with different quantities of paramagnetic material to determine how much broadening could be tolerated. Stefan Berger and Siegmar Braun suggest using 0.1M Cr AcAc in "200 and More NMR EXperiments". <br /><br />GlennGlenn Faceyhttps://www.blogger.com/profile/05146575170575279335noreply@blogger.comtag:blogger.com,1999:blog-3300702123878659843.post-85012709669784596662018-04-04T11:09:24.920-04:002018-04-04T11:09:24.920-04:00Dear Glenn,
How does the paramagnetic agent do fo...Dear Glenn,<br /><br />How does the paramagnetic agent do for T2 relaxation? Do you worry about line broadening? Thanks!<br /><br />Bo ZhangAnonymoushttps://www.blogger.com/profile/18318763358214260920noreply@blogger.comtag:blogger.com,1999:blog-3300702123878659843.post-89032985387005217982018-02-26T15:12:49.750-05:002018-02-26T15:12:49.750-05:00VNM,
Aside from the natural abundance, one must al...VNM,<br />Aside from the natural abundance, one must also consider the gyromagnetic ratio, spin number, relaxation properties and line widths when considering sensitivity.<br />GlennGlenn Faceyhttps://www.blogger.com/profile/05146575170575279335noreply@blogger.comtag:blogger.com,1999:blog-3300702123878659843.post-69987161491372741692018-02-25T00:52:54.749-05:002018-02-25T00:52:54.749-05:00Dear Glenn,
Relatively compare, 1H nuclei is...Dear Glenn,<br /><br /> Relatively compare, 1H nuclei is more sensitive than other nuclei as it is more abundant nucei. What are the other factors that affect sensitivity aspects, except natural abundance ? For example, some nuclei are less sensitive although their natural abundance appreciably good ? Please clarify my doubt regarding this aspect..<br /><br />With thanks<br />VNM<br />Anonymousnoreply@blogger.comtag:blogger.com,1999:blog-3300702123878659843.post-68419270233362281202016-09-26T08:58:31.382-04:002016-09-26T08:58:31.382-04:00Anonymous,
The relaxation rate for a spin, 1/2 nuc...Anonymous,<br />The relaxation rate for a spin, 1/2 nuclide influenced by a paramagnetic agent depends on the the square of the local field which in tern depends on the total spin, S and the reciprocal 6th power of the average distance between the unpaired electron and the spin. I would not expect this term to vary much with temperature. The relaxation rate also depends on a correlation function which in tern depends on the correlation time in a Lorentzian manner. As the temperature is increased, the correlation time becomes shorter. For molecules in the extreme narrowing limit (small molecules) this means that the relaxation becomes less efficient and the T1 relaxation time therefore gets longer.<br /><br />GlennGlenn Faceyhttps://www.blogger.com/profile/05146575170575279335noreply@blogger.comtag:blogger.com,1999:blog-3300702123878659843.post-23069696563770365722016-09-24T10:58:42.031-04:002016-09-24T10:58:42.031-04:00Dear Glenn, what is the effect of heating the NMR ...Dear Glenn, what is the effect of heating the NMR sample containing the compound of interest and Cr(acac)3 as the relaxation agent?Anonymousnoreply@blogger.comtag:blogger.com,1999:blog-3300702123878659843.post-44270386498248977412013-01-22T09:04:31.865-05:002013-01-22T09:04:31.865-05:00Anonymous,
Thank you for your question. The NOE&...Anonymous,<br /><br />Thank you for your question. The NOE's decay at the 13C T1 therefore you should leave at least 5*T1 without decoupling between scans. If you are willing to put a bit of paramagnetic relaxation agent in your sample, you will shorten the T1 and also reduce the dipolar contribution to the relaxation allowing you to use shorter delays and get more quantitative results.<br /><br />GlennGlenn Faceyhttps://www.blogger.com/profile/05146575170575279335noreply@blogger.comtag:blogger.com,1999:blog-3300702123878659843.post-42238874663891773242013-01-21T16:37:32.136-05:002013-01-21T16:37:32.136-05:00Hi Glenn,
My sample contain all protonated carbon...Hi Glenn,<br /><br />My sample contain all protonated carbons. the T1 for protonated carbons are about 2.5 seconds. Is there a relationship between T1 of the carbon and the length of the period not to decouple (for NOE to dissipate) in order to be quantitative?Anonymousnoreply@blogger.comtag:blogger.com,1999:blog-3300702123878659843.post-17376864306945826772013-01-03T09:30:45.298-05:002013-01-03T09:30:45.298-05:00Anna,
Thank you for your question. Yes it is pos...Anna,<br /><br />Thank you for your question. Yes it is possible in the same way it is possible with a standard broadband probe as long as the T1's and NOE's are respected. The only difference is that you will have less sensitivity compared to a standard broadband probe due to the smaller filling factor.<br /><br />GlennGlenn Faceyhttps://www.blogger.com/profile/05146575170575279335noreply@blogger.comtag:blogger.com,1999:blog-3300702123878659843.post-42635781513714809732013-01-03T08:15:39.491-05:002013-01-03T08:15:39.491-05:00Hi Glenn do you think it is possible to get a quan...Hi Glenn do you think it is possible to get a quantitative 13C spectrum from by a BBI probe (internal coil for 1H snd external coil for 13C)?<br /><br />AnnaAnonymousnoreply@blogger.comtag:blogger.com,1999:blog-3300702123878659843.post-57192415724743845002008-07-21T09:02:00.000-04:002008-07-21T09:02:00.000-04:00John,According to Stefan Berger and Siegmar Braun ...John,<BR/><BR/>According to Stefan Berger and Siegmar Braun in "2oo and More NMR Experiments" p.318, the appropriate amount of Cr(acac)3 to use is 0.1M or 35 mg in 1 ml of solution.<BR/><BR/>GlennGlenn Faceyhttps://www.blogger.com/profile/05146575170575279335noreply@blogger.comtag:blogger.com,1999:blog-3300702123878659843.post-53434945673467127422008-07-19T11:34:00.000-04:002008-07-19T11:34:00.000-04:00Glenn,Is there a rule-of-thumb for the molar quant...Glenn,<BR/>Is there a rule-of-thumb for the molar quantity of paramagnetic reagent required? If not, how would you suggest one optimise the reagent use - add 5 mg and run a 1-scan acquisition, repeating as necessary?Anonymousnoreply@blogger.com